Worked DI-2008 slist example
I figured out my problem in writing up a question.  Here is the corrected write-up for posterity.
To understand the slist command, decode the example given for a DI-2008 (p. 40):
slist 0 2562
slist 1 2564
slist 2 3078
slist 3 9
slist 4 10
slist 5 8
Here is how to calculate slist 2 3078.
The slist command has form: slist <zero-offset> <16-bit config word>
The <zero-offset> determines the sampling order.  The <16-bit config
word> defines the function (i.e. input/rate/count) and
scale/thermocouple.
For "slist 2 3078", the offset is 2.  This means it will be sampled
third (b/c of zero-based indexing).  3078 is base-ten.  This
is 0000 1010 0000 0010 in binary.
The <16-bit config word> parameter works as follows.  The 16 bit
positions are grouped and each group has a different meaning.  The
meaning of each group is provided by a protocol table.  The position
numbering of bits goes right-to-left.  For 3078 = 0000 1010 0000 0010
the positions are:
bit: (  0   0   0 ) (  0   1 )  ( 1  0  0 )  ( 0  0  0  0 )  ( 0  1  1  0 )
pos: ( 15  14  13 ) ( 12  11 )  (10  9  8 )  ( 7  6  5  4 )  ( 3  2  1  0 )
The meaning of each group is as follows.  Group 3-2-1-0 represents the
four right-most columns of the DI-2008 Scan List Word Definitions
table (p. 42).  Following the row to the left-most column gives the
corresponding function (Analog Channel/Digital In/Rate/Count).  Group
7-6-5-4 is not used.  The table only states that "unused bits = 0".
Group 10-9-8 gives the ROW to use in the DI-2008 Measurement table
(p. 43).  Group 12-11 is the COLUMN to use in the DI-2008 Measurement
table.  Position 12 is mode, position 11 is range.  Group 15-14-13
also seems to be unused.  Again, the table only says, "unused bits =
0".
Decoding goes as follows.
Group 3-2-1-0 is 0110.  This is the 6th from last row on the far right
of the DI-2008 Scan List Word Definitions table (p. 42).  The
left-most column of this row is "Analog channel 6".
Group 7-6-5-4 is all 0's, so unused. Ignore.
Group 10-9-8 is 100.  So, use the fifth ROW of DI-2008
Measurement table (p. 43).
Group 12-11 is 01.  The mode is 0 and range is 1.  Use the second
column of the DI-2008 Measurement table (p. 43).  Since Group 10-9-8
is 100, this corresponds to "+/-2.5V".
Group 15-14-13 is all 0's, so unused. Ignore.
Thus, 3078 = 0000 1010 0000 0010 => Analog Channel 6 +/-2.5V.